Forums › BB Series Discussions › "Max Input" / "Ref Offset" parameters understanding for EMI precompliance › Reply To: "Max Input" / "Ref Offset" parameters understanding for EMI precompliance
Andrew
Moderator
Vincent,
When the ref offset is zero, the CW peak you measure in the software should be equal to what is on the input port. It sounds like this is the case for you based on your first message? Changing the max input should not affect the CW peak reading (assuming you only use values above the input level). You can contact me directly at aj@signalhound.com if you have further questions.
Regards,
Andrew