Forums › General Discussions › Phase noise – result exporting › Reply To: Phase noise – result exporting
hendorog
Participant
I think the recent fix for this PN export issue still has a problem:
The frequency offset values in first column of the exported CSV are all a factor of 10e4 too small if the sweep is from 1kHz -> 1MHz.
If the sweep is from 100Hz -> 1MHz they are all a factor of 10e5 too small.
If the sweep is from 10Hz -> 1MHz they are all a factor of 10e6 too small.