Forums › BB Series Discussions › What is wrong with my measurments › Reply To: What is wrong with my measurments
Justin Crooks
Moderator
This is a difficult question to answer. My first question is, what antenna distance are you using? Is it the same distance in both cases (certified test and pre-compliance test)? If not, there is a 20 log (d2/d1) dB correction to apply.
Obviously your chamber is not anechoic, which will change results, and VSWR on your antenna and amplifier can change things. I can see these stacking to produce maybe a 6-8 dB variation, but 20 dB seems a bit excessive.
Do you have two of the antennas? If so, you could calibrate your system in position using a signal generator instead of your DUT…