Forums › SA Series Discussions › Is SA124B suitable for EMC pre-compliance testing? › Reply To: Is SA124B suitable for EMC pre-compliance testing?
Andrew
Moderator
Hi ghuxtable,
Yes, changes in the reference level affect sensitivity, which change gain and attenuation of the instrument. A lower reference level implies you will be making measurements on smaller signals, in which we increase the sensitivity.
In standard sweep mode, linear scale frequency is the only scale possible. There is no way to plot on a logarithmic scale outside of the precompliance mode. I apologize for the inconvenience.
Regards